编辑: jingluoshutong | 2018-07-31 |
154 Sm/147 Sm). Because of the high precision of the measurements, the error on Nd and Sm concentrations is only carried by the error on the composition of the spike, which is better than 0.5% (2σ). Results are presented in Supplementary Table 1. The unspiked Nd cuts for high-precision
142 Nd/144 Nd measurements were performed on the 9-faraday cup Triton thermal ionization mass spectrometer at the Johnson Space Center (Supplementary Table 2). The analytical protocol described by Caro et al.27 was employed, where Nd isotopes are measured in multidynamic mode with rotating amplifiers between blocks. The Nd isotopes were measured as positive metal ions. All measured data are normalized to
146 Nd/144 Nd = 0.7219 to correct for instrument mass fractionation during the runs. The interference of
142 Ce on
142 Nd was monitored with
140 Ce. Corrections for this interference to
142 Nd/144 Nd ranged from 0.2 to 9.5 ppm, and for all but one AMES run and
4 Iceland picrite runs, ≤
4 ppm and within 2σ internal and established 2σ external precisions on
142 Nd/144 Nd (Supplementary Table 2). The external precision is expressed as
2 standard deviations (2σ): 2σ =
2 * x ? x ( )
2 ∑ n ?
1 ( ) Samarium interferences on the Nd isotopes were monitored with
147 Sm, which was never detectable on the faraday cups above background noise. Scans on the electron multiplier showed that
147 Sm was never above
1000 counts/sec, and therefore negligible. The amount of Nd extracted in the SNC samples measured ranged from approximately
300 to
600 ng. In order to constrain the accuracy and external precisions for the measured SNC data, standards were run with similar concentrations. A series of runs for two standards, an AMES in-house internal standard and LaJolla, are shown in supplementary Figure 1. Internal precisions of ≤ ±3 ppm were achieved for
600 ng standard aliquots for measuring times of
4 to
6 hours (Standard Set 1, Supplementary Table 2). The signal intensities were
4 to
6 V on the
142 Nd beam. The average for these standards was
142 Nd/144 Nd = 1.1418402±0.0000034 (n=14, 2σ) and is used as the reference value of ε142 Nd =
0 for calculating ε142 Nd of the standards and unknowns. The 2σ external precision for these
14 standard runs for ε142 Nd is ±0.03 (±3 ppm). The average
143 Nd/144 Nd values for Ames standards (n=10) and La Jolla standards (n=4) give 0.5121359±0.0000026 (2σ) and 0.5118491±0.0000007 (2σ) respectively (Supplementary Table 2). A set of
142 Nd-enriched standards were measured in order to check for accuracy over a range of
142 Nd/144 Nd. The enriched standards were prepared by Chi-Yu Shih at the Johnson Space Center by adding >
99% pure
142 Nd spike to the in-house AMES Nd standard in measured quantities. All the data are presented in Supplementary Table 2. 1.141820 1.141840 1.141860 1.141880 1.141900 1.141920
142 Nd/144 Nd AMES La Jolla Std. Avg. 0.218 AMES 0.430 AMES 0.638 AMES ±3 ppm 2σ 0.218 AMES 0.430 AMES 0.638 AMES ±3 ppm 2σ 1:1 ε142 Nd - Gravimetric
0 0.1 0.2 0.3 0.4 0.5 0.6 0.7 0.8
0 0.1 0.2 0.3 0.4 0.5 0.6 0.7 0.8 ε142 Nd - Measured Supplementary Figure 1. High-precision measurements of standards for
142 Nd/144 Nd. Error bars are ±2σ. As shown in Supplementary Fig.
1 and Supplementary Table 2, for gravimetrically prepared
142 Nd-enriched standards with ε142 Nd at 0.218, 0.430, and 0.638, the measured ratios for these enriched standards fall on a 1:1 line with the values obtained gravimetrically (Supplementary Fig. 1). This demonstrates that the accuracy of these measurements are reproducible within this external precision of ±3 ppm (2σ) for materials that vary over a range of >